We are a training core, so once a user is trained, they are free to use (most) instrumentation at will. Although, as you can imagine different users have differing skill sets. Thanks for the tip on the 96 cap array - we wouldn't be using that - too much throughput for us, but good to know.
------------------------------
Andor Kiss - Center for Bioinformatics & Functional Genomics
Director
Miami University
Oxford OH
------------------------------
Original Message:
Sent: 11-13-2025 11:39
From: Daniella Wilmot
Subject: Agilent Fragment Analyzer - Does it really need an SC?
Our lab's FA has always been very robust. We have a service contract, but have not needed it for repairs - we replace the capillary ourselves but get an annual PM where the instrument is cleaned and the fluidics tubing is replaced. Issues we've had have been related to samples or reagents.
We do restrict its use to trained operators (it's not a shared instrument) and adhere to cleaning and maintenance schedules. We also found that the 96-capillary array did not work for us - it was too easy for samples over the assay concentration range to bleed signal into surrounding capillaries. The 48-cap array works fine for our throughput.
One caveat: our instrument is older and was manufactured by the original company that was later purchased by Agilent (Advanced Analytical). I don't know if there have been changes to the design/manufacture since then.
------------------------------
Daniella Wilmot
Research Associate IV
UMass Chan PacBio Core
Shrewsbury MA
Original Message:
Sent: 11-11-2025 05:59
From: Andor Kiss
Subject: Agilent Fragment Analyzer - Does it really need an SC?
Considering buying an Agilent FA 5200 to QC nucleic acids, need to be able to size large PacBio libraries. Doe I really need an SC or is this instrument pretty stable and worry free?
------------------------------
Andor Kiss - Center for Bioinformatics & Functional Genomics
Director
Miami University
Oxford OH
------------------------------