Hi Jodie,
I'm assuming you can load the loop with the syringe and some liquid actually comes off of one of the drain connections (see
http://www.hplc.ru/catalogs/hplc_7725-9725_injectors.pdf for drawings). Then none of the tubing is clogged, so it must be one of the channels in the rotor/stator (I don't know which part has the channel grooves).
If sonicating all the parts does not help, I'd suggest connecting it in the opposite direction (swapping pump and column connections) and trying to push some liquid through.
hope this helps.
zdenek
------------------------------
Zdenek Svindrych
Research Scientist
Dartmouth College
Lebanon NH
------------------------------
Original Message:
Sent: 01-30-2023 09:56
From: Jodie Franklin
Subject: Waters 7725i HPLC Injector
good morning, I am using a Waters 1525 pump with a 7725i manual injector. When the injector is in "inject" mode, everything flows correctly. When we put it into "load" mode the flow stops and the pressure goes up until it faults. I have changed the rotor seal and the stator face and I still have the same problem. I also checked the flow into each position of the stator and they seem to be okay. Does anyone have any ideas on where to go next before I change the entire manual injector? thanks! Jodie
------------------------------
Jodie Franklin
Research Associate
Johns Hopkins Univ/Sch of Med
Baltimore MD
------------------------------